BSA as a high mass calibrant for MALDI?

Author:   Mark Hail  
Posted: 2/23/01; 10:19:43 AM
Topic: BSA as a high mass calibrant for MALDI?
Msg #: 153 (top msg in thread)
Prev/Next: 151/154
Reads: 33105

There has been a lot of discussion recently in the ABRF discussion group about high mass calibrants for MALDI.  It seems that BSA is used by a lot of MALDI practitioners.  As a few have pointed out, it's probably not a very good calibrant.  The data below was obtained by ESI/MS.  As you can see, BSA is quite heterogeneous.  This sample was even put through a desalting step with an RP-trap cartridge.  Most likely, you will not be able to resolve this heterogeneity by MALDI, so the mass you would observe would represent a composite of all of these peaks plus any additional matrix adducts.

BSA ESI spectrum: